I bought an ostensibly "good working" HP 3468A Bench multimeter for only $84.
It hasn't arrived but the seller did a few tests with it and says it is working.. plus it passes the "self test" mode too.
The meter has no internal adjustments and gets calibrated via software changes via front panel or remotely via RS232 port.
The original factory calibration offsets are stored in CMOS using a 2.9v .75mah Lithium battery. This battery is supposed to maintain values up to 10 years.
This meter was made about 1994 or so I think.. so that battery may be near end of life.
I don't want to pull it to replace it and possibly loose my settings ( then I'd have to send it out to be calibrated maybe $$)
I bought another 3volt lithium .8maH battery to replace it with.
How do you suggest I replace it live with out screwing anything up.
There is a diode -->|-- leading to the CMOS from the plus end and a .1uf with one end to ground...
.. that may hold up CMOS during the change but not sure that is really what it is there for and how long it will work while I make the swap.
Do you think that .1uf would remain charged long enough?
My plan is to try to somehow parallel in a new one then remove the old... but just how?
Any ideas?
|
Nice upload of the schematic. I see the mystery of the 0.1 cap, but as I'm sure you've suspected it likely suppresses the Vcc. I would solder in the + and - leads and connect up your new battery.
:
I put two AA cells in series with a switching diode and first hooked that as a temp source to the circuit to supply 3v.
Then I clipped out the old 3v battery and soldered in the new 3v lithium.
It all seems to be working fine.
I guess I got that old battery out in the nick of time... as it has practically no juice left at all.
I connected a 2.5v 500ma lamp across it and it would not light at all.... however the two AA's in series lit it up just fine.
So I guess that old lithium was really weak.
It has a name:
Eternacell sdx
3volts No.9511
hp part no. 1420-0278
I replaced it with a 3 volt 800mah lithium.