Sir David. . . .
Well here we are again exploring yet another branch on the total repair aspect. Wayyy back
on day one, was not my first few "breaths " from my keyboard being relevant to the confirmantion
of the presence and tolerances of ALL of the LVPS's developed supply voltages?
That, even being with a very detailed explanation of all of the supply voltages, and the
dependent fact that almost all were "electrically interlocked" to each other
That which you are running up against now is the fact that the hoz positioning, is
related to a capability of needing a completely seamless shift of a DC voltage. Now that
might be dependent, at one point, of a swing down from being a +DC voltage and down to
a zero voltage axis and then a swing onto a transitional shift into being a - voltage
level.
Remember, this scope extensively uses DC coupled differential stages which are dependent upon
use of bipolar supplies.
Alas, your -150 supply is a bit higher than its prescribed level so that is upsetting other conditons, since it is used in so many other portions of the set.
ADJUNCT ASIDE:
I don't remember the units hoz positioning control, being other than the common dual
concentric control, that is using a high value pot along with a smaller pot such that the combo
gives a coarse and vernier adjustment capability.
You then just place the start of trace, at the very left graticule indexing of the grid display matrix.
Now you ALSO have a Helipot control on the front panel in the form of the vernier delay adjustment
of the B mode time delay onset, if using the "B" delayed sweep functions capability of the unit.
It would NOT be relevant to you at all, until you get this unit percolating, you will be placing
the HOZ sweep option switch at its very top, upper,center "A" sweep mode option.
Let me now go into a detail of the basic operation principle of the - 150 supply.
Seems like that I tacked the LVPS Supply Working Schematic on to many different points, but
this one seems to kick right in . . .unless on rural- party line - hand crank- dial up . . . .meet
you here, same time on Thursday !
Seek out:
http://www.nostalgiaair.org/Forums/Messages/530/M0080530.htm
Looking at the -150 supply portion, and reiterating again, the - supply line is a direct
connection from - line of the DC raw supply, with no inline obstructions, with it then emerging
at the far right, as being the designated -150 supply terminus.
And NOW, the positive portion of the raw DC supply, is then feeding on over to the plates of
a triad bank of (3) 12B4 power reg tubes.
The basic regulation concept of this stage is dependent upon that trio of tubes experiencing
an initial turn on with their 1st grids conduction biasing being acquired via that R729 which
you see going up to its supply source . . .the +110 regulated supply source, so
now I ask you . . . . GOT MILK ? . .errrr make that +100 VDC supply voltage ?
As it stands, if that bias voltage is present and is unhampered, the 12B4 trio would be driven
into so full of a degree of conduction, that there would only be the voltage drop across the
plate-cathode of the tubes that is making a path to ground via the cathodes.
In that conditon, the supply would be having a MUCH-MUCH greater output voltage than the desired
-150 VDC benchmark.
( . .BTW . . .Now, isn't that the EXACT condition which you are now experiencing ?)
Soooooo go back to the schema and notice the daisy chaining of all of the 12B4 1st grids, along with
their final emergence being over to the left and downwards to the plate of the V700 comparator stage
where,in actuality, it is then providing that stages plate supply voltage also.
Then, there is a straight forward technique in the screen voltage acquisition for that V700 stage.
When looking at its 1st grid control voltage, we see it as being routed in from a tap off
of a resistive voltage divider(R707-706) that is going over to connect across the plate of a DC
differential pair- DC coupled amplifier stage.
That same plate is receiving ITS supply voltage via the R719 resistor from the + 100 VDC supply.
The idea now in the regulation aspect, is for that 12B4 trio to not FULLY conduct, but to only
partially conduct, such that they can be responsible for creating the voltage drop which decreases that initial FULL raw DC supply, on down to the required -150 volt level.
That is being accomplished by two conditions, initially by the aquisition of a stable voltage reference source to input at the 1st grid of the left triode of that V712 differential DC amp stage.
Notice the presence of the 5651 gas tube regulator, being placed between the negative supply buss and a current limiting resistor R710, a complete series power loop is then made by the final connection of the other electrode of the 5651 to ground.
There is then the series resistor isolation that is provided by resistor R711 and the final presence of a slightly dropped initial ~87 voltage level that was developed across the conducting 5651 gaseous regulator reference.
The next aspect then would be the right half triode of V712 where its plate is receiving its supply from the R719 resistor.
Look now at the R718-R716-R715 voltage divider trio, with their interrelated RATIO of values, being chosen such that a min and max of adjustment range limits could be acquired.
That sampling across the -105 VDC output voltage , is fed on in thru isolative resistor R717 and then
presents a sampled volage to the 1st grid of the right triode of the diferential pair.
Note that the pair of triodes cathodes are tied together and then sharing a common cathode resistor.
Look and see the fixed voltage reference over at the left triodes first grid.
Also see the voltage dividers sample of the level of voltage present across the -150VDC
supply that is feeding into the right triodes 1st grid.
With that DC level sharing of the stages, that voltage level present at the #6 plate of the right triode will walk up and down in value, with any shift of that -150 supply buss level.
A portion of that difference is tapped off via the R707-R706 resistive voltage divider to then feed
in as a controlling and corrective grid bias, to the 1st grid of the V700 comparator stage.
If there is a need for a voltage correction, from there being too high of a voltage presence at the
-150 VDC buss sampling, the slight presence of a more positive voltage to the 1st grid of V700 then
increases its conduction.
More conduction in that V700 stage then means that the common V700 plate to 1st grid connection to
the 12B4 trio is going to pull down on, and bleed down that grid turn on + level that is continually
being provided via R729 from the +100 V supply.
The end result is the soaking up of the excess voltage via the 12B4 cluster and the regulation of the output buss down to the prescribed -150VDC.
Regulation Technique Summation:
The left V712 triode's 1st grid receives a sample of the stable voltage reference.
The right V712 triode's 1st grid receives a sample of a segment of of -105 V supply buss level.
They compare . . . and any variance from the norm will result in a shifting up or down of the plate
supply voltage of the right triode of V712.
That potentially variant voltage then feeds over to the 1st grid of the V700 comparator stage, its
potentially variant grid biasing swing could then provide for more or less conduction in accordance
as to whether it was making a + swing or a - swing.
That varying degree of conduction at the plate of V700 IS ALSO directly tied into the 1st grids of the 12B4 triodes.
Net Result . . . . would then be:
If V700 conducts more heavily, it pulls down the normal + grid bias present on those 12B4 stages via the R729 supply resistor from the +100VDC supply.
As they conduct less, that trio starts absorbing more of the surplus voltage presence, thus the -105 buss has a reduced voltage level presence, with it then shifting down to the specified -105 level.
Same control effect on the other situation of the -105 supply output being at LESS that its desired output voltage level.
In that situation the V700 would be needing to conduct LESS in order for there to be less pull down of the static + bias being fed into the 12B4's 1st grids.
How to do-it-to-it:
DVM probe to chassis ground, and then you use its other probe to 2-4 or 7 of the 5651 . . if you then find an ~87 V presence, the gas tube is firing and regulating as it should .
Move metering to the center terminal of R716 pot [ -150 supply adjust ] and swing the pot thru its range, to confirm an according voltage swing on the meter. . .if so.
Move metering to the Pin #6 plate of V712, keep up the same swinging of the R716 thru its range to see if there is a like variable response in the swinging of that plate level.. . .if so.
Move metering to the Pin #1 1 st grid of the V700 comparator stage plate to see if there is a like variable response in the swinging of that 1st grid voltage .. . .if so.
Move metering to the Pin #7 of V725 1st grid to see if there is a like variable response in the swinging of that 1st grid voltage .. . .if so.
NOW, you MOVE metering over to being across Pin #1 and pin #9 of V725 to see what the maximum voltage developed is, at the setting of the -150 adjustment pot which results in the highest voltage reading.
If that is not high enough, that is why your raw B+ level is not being able to be adjusted down to that -105 standard level.
Go back and check the screen, plate, first grid and cathode voltages of the V700 and check for their variance from the schematics estimation of those voltages.
ZUJ'ing
73's de Edd