Thanks
http://docs.google.com/gview?a=v&q=cache:VZZxQNeZybAJ:www.combatindex.com/store/tech_man/Sample/Oscilloscopes/TB_11-6600-254-50.pdf+hum+balance+control+oscilloscope&hl=en&gl=us
:Maybe this link will help.
:
:http://docs.google.com/gview?a=v&q=cache:VZZxQNeZybAJ:www.combatindex.com/store/tech_man/Sample/Oscilloscopes/TB_11-6600-254-50.pdf+hum+balance+control+oscilloscope&hl=en&gl=us
:
http://bama.edebris.com/manuals/dumont/304
Carl T
:That article is probably what I need, however only thr first 3 pages are there I need all of them.
:
:
::Maybe this link will help.
::
::http://docs.google.com/gview?a=v&q=cache:VZZxQNeZybAJ:www.combatindex.com/store/tech_man/Sample/Oscilloscopes/TB_11-6600-254-50.pdf+hum+balance+control+oscilloscope&hl=en&gl=us
::
:
When the y input is grounded there is no ripple. Could I boost the values of one of the filter caps, if so which one? there are 5.
Thanks for your help
Ian,
:Have you looked at the manual on BAMA?
:I believe it will help you locate the hum control.
:
:http://bama.edebris.com/manuals/dumont/304
:
:Carl T
:
::That article is probably what I need, however only thr first 3 pages are there I need all of them.
::
::
:::Maybe this link will help.
:::
:::http://docs.google.com/gview?a=v&q=cache:VZZxQNeZybAJ:www.combatindex.com/store/tech_man/Sample/Oscilloscopes/TB_11-6600-254-50.pdf+hum+balance+control+oscilloscope&hl=en&gl=us
:::
::
:
On all voltage setting there is ripple, on .1 volts it is over 2 inches , on 100v it is .2 inches.
As for the isolation transformer, I have never repaired a AA5 or hot chassis radio. All of mine are prewar.
:When you ground the input, then showing no ripple, your power supply is working the way it should. What voltage setting are you using? if it's on the lowest setting, it will pick up noise right out of the air. Might check the probe you are using too. The shield ( ground side ) does need to be going to a good ground connection at the probe end. Better be using an isolation transformer too on any of your radio projects.
:
|
Dumont Scope Schematic Mark-up: ![]() |
|
Dumont Scope Schematic Mark-up: ![]() |
|
Dumont Scope Schematic Mark-up: ![]() |
![]()
On that scope, I thought that I used to have the same unit hard cable connected up, for using as the display to a bench set up for the sweep alignment of 70 mhz radar IF strips, but, no, I see now that I was using a Whoolet Plastered 130 series. Sounds like to me that you ought to check out the filtering on the power supply. With that high level pollution of the trace, sounds like its early in the stages of the Y / Vert amplifier section. Consult the supplied schematic blow-up and note the initial input into the vertical amp section via the red circled BNC (or possubly an older SO) connector.
The input signal is then differentated by the use of the input switch such as to whether to be in either the AC input mode or the straight DC mode via the switched in coupling capacitor.[Blue-green markup] The signal is then processed thru cascaded sets of 5 differential amps in the form of V1-V4 pairs and then a final deflection plate driver pair in the form of V5/6. Now, take note of the markup of the B+ supply levels to those stages, with the highest buss from the buss deriving power from the 5Y3. . . . [RED BUSS] Then the sub B+ level is being marked up as the [FUSCIA], that is the supply for the lower level amps, with the supply being derived and dropped from the main buss just mentioned. And yes, the sets OBie Kanobie 2 gas regulator tube on that line is working , since you see its glow from the top . . .(since it is not silvered up inside the envelope and you can ACTUALLY see it).
Be sure if it . . . . and a bit less important, is the earlier main B+ filtering at the 5Y3 at its [RED MARKUP . .1 and 2 squares] being in the form of C39 and 40. Another thought would be the possibility of fil- cathode leakage in the V1 thru 3 preamp tube stages, but I think that it would have made your quiescent trace positioning a bit squirrely or off of scale. Swap tubes V1-4's positioning and see if your problem changes in amplitude, . . . . or even abates a bit. I just additionally enhanced other filters used in the set, but the primary ones of suspicion would be the one on the [FUSCIA] and [RED] lined markups. As far as "calibration" on the unit and with it being a fiffy cent item , but that being well under the typical $20 evaluation for that unit. Yet this unit is head over heels better than an EICO or HEATHKIT of that vintage.
Soooo what you do now is position the trace to the bottom line graticule of the gridwork and then put the vert input sensitivity to 1 V per CM switch cal setting.Set the input coupling mode to DC, so that a mere DC voltage level can be used as your cal source.
Same procedure for the other decades, but using correspondingly different reference voltages. The inherent trace width of a scope line, precludes ascertaining any REAL accuracy on readings anyhow. Lastly, as per the "only using on old radios", this unit, with its limited 100's of KC's bandwidth would only suffice for that type of service or for audio display . . . .or as a display on a sweep alignment setup.
![]() |