The meter is ok, full scale at 1mA. I double-checked all switches and they are ok. The problem is not related to a specific socket or tube type. They all read low. Unfortunately I only got the tube chart with the tester. If someone has the schematics or calibration instructions I would love to get a copy.
Thanks,
Wilhelm F. Schroeder
Paderborn, Germany
You can put a milliammeter in series with the plate, and see what the current is. There's no guarantee that the B&K and the Jackson apply the same plate voltage however.
I mentioned the result from the I-177 and Jackson only to make it clear that the tube is fine.
The point is: if the 606 consistently reads around 80% instead of 100% for different types of known good tubes then there must be something wrong with it. That's why I am looking for schematics and the procedure to check/set calibration. Do you have a 606 in your collection?
Wilhelm
: The meter is ok, full scale at 1mA. I double-checked all switches and they are ok. The problem is not related to a specific socket or tube type. They all read low. Unfortunately I only got the tube chart with the tester. If someone has the schematics or calibration instructions I would love to get a copy.
: Thanks,
: Wilhelm F. Schroeder
: Paderborn, Germany
I have a Lafayette emission tube tester, and the instruction manual mentions directly that the meter circuitry was designed so that no tube would ever reach full reading on the scale. The B&K may have been set up in a similar manner. My $0.02 :-)